applied-measurement · 24 min read

More Digits Do Not Make a Better Garden Thermometer

!Two differently shaped unbranded thermometers in a shaded field calibration rack, display faces turned away. (generated illustration)

Two differently shaped unbranded thermometers in a shaded field calibration rack, display faces turned away. (generated illustration)
Two differently shaped unbranded thermometers in a shaded field calibration rack, display faces turned away. (generated illustration)
Readings from a reference at 20 degrees Celsius (illustrative; data table below)
Readings from a reference at 20 degrees Celsius (illustrative; data table below)
CategoryValue (degrees Celsius)
Fine display 120.84
Fine display 220.87
Coarse display 120.2
Coarse display 220.1

Illustrative only, not taken from a source dataset. Illustrative readings show finer display with greater observed deviation in this example only.

A new thermometer arrives with two places after the decimal point. The older instrument displays only one. The natural instinct is to put the new one in charge of the trial: surely a screen that reports hundredths of a degree knows more than a screen that reports tenths. Then both instruments are compared with the same reference near 20 degrees Celsius. The fancy display tells a consistent story more than eight tenths of a degree above the reference. The older one lands much closer. The extra digits did not disappear when the instrument was wrong. They described its indication in finer steps.

That does not mean a finer display is bad. It can be useful when the sensor, reference, readout and operating method support a reliable measurement at that scale. The mistake is treating the visible increment as a credential for the rest of the instrument. Garden measurements are especially vulnerable to this shortcut because conditions are easy to describe casually and hard to reproduce precisely. A thermometer beside a tomato leaf, a probe in shaded soil, and a weather sensor protected from sun can all be working as designed while reporting different quantities in the same general location.

A better question than “how many digits are on it?” is “which temperature do I need, and how close can this procedure get to it under the conditions that matter?” To answer that, separate four properties: resolution, repeatability, observed deviation from an appropriate reference, and uncertainty. Then check that the test represents the way you intend to make decisions. A small, documented comparison is more revealing than a large tally of digits or an impressive specification read out of context.

Read the chart as an invented check, not a product comparison

The four bars compare two hypothetical indications from each thermometer against an assigned reference temperature of 20 degrees Celsius. The fine display reads 20.84 and 20.87 degrees Celsius. The coarse display reads 20.2 and 20.1. These are constructed values, not results from the NIST Technical Note 1297 linked with the chart, and they are not an instrument rating. The chart gives the indications themselves. To find deviations for this example, subtract the assigned reference temperature from each reading.

The fine display's deviations are plus 0.84 and plus 0.87 degrees Celsius; those of the coarse display are plus 0.2 and plus 0.1 degrees Celsius. The fine pair spans 0.03 degrees Celsius and the coarse pair spans 0.1 degrees Celsius. Computing the means from the shown values gives 20.855 and 20.15 degrees Celsius respectively, or illustrative mean deviations of plus 0.855 and plus 0.15. Those mean digits are arithmetic, not proof the instruments achieved that reporting resolution. With two readings per device, it would also be a mistake to treat the example spans as established long-term repeatability.

The four values give a counterexample to the inference that a finer display always produces a closer reading. The numbers do not prove a general rule that coarse thermometers are better, that the fine device is wrong by the same amount at all temperatures, or that the reference is exact in a real comparison. In an actual test the assigned value of the reference has its own uncertainty. Two probes may be exposed to different temperatures if their positions or timing differ. The example assumes those sources of difference are not in play so the distinction between the displayed increment and observed deviation is easy to see.

The fact that all four results are positive deviations is worth noticing. Not even the closer device exactly matched the hypothetical reference on the two shown checks. Calling one “accurate” without naming the use could conceal this. A useful statement is conditional: in this example the coarse device's two indications were closer to the assigned temperature than the fine device's two indications. To determine whether either is suitable, specify the temperature range, acceptable decision risk and measurement method. A notebook intended to compare broad daytime warming can tolerate a different discrepancy from a close study of two nearly identical sheltered beds.

Resolution describes the steps, not the temperature

Resolution is about distinguishing a small change in an indication. An obvious feature of a digital display is its displayed increment. A display advancing from 20.84 to 20.85 in the mode used by the observer shows an increment of one hundredth of a degree Celsius in that range. It might be displaying a rounded representation of a more detailed sensor stream, or it might be presenting values produced by a filter or model. The last digit alone does not tell you how much independent physical information is present. A coarser readout can merge real variations into one value; a finer one can show harmless fluctuations or a stable discrepancy with great visual confidence.

Imagine a display that rounds to the nearest tenth. Two distinct unrounded values within the same rounding interval can be shown as 20.2. The unchanged display does not prove that the sensor experienced exactly the same temperature. Conversely, a shift near a rounding boundary can take the reading from 20.2 to 20.3 even when the underlying change is considerably smaller than a tenth. This is why changes between adjacent display values are not uniformly interpretable as physical changes of the same size. Averaging rounded numbers cannot recreate the individual unrounded observations that the instrument never stored or exported.

Some devices offer a display setting and a separate export setting. An app can also change the number of digits shown without changing the sensor or its error. Check the physical unit, logger configuration, time-averaging method and data export precision instead of assuming they follow the printed screen. If the display reports Celsius to a tenth and an export formula converts it to Fahrenheit with several decimals, the calculation may have more digits without having further measured detail. Good reporting adapts the precision to the evidence and the decision, while retaining the original file for reproducibility.

One practical test of acceptable resolution begins at the scale of a real question. If a study intends to distinguish a change of about one tenth of a degree, a display that only updates once per degree provides little direct leverage from ordinary individual readings. But a display that updates to one thousandth is not automatically adequate. Check how much noise is present, how stable the sensor is, how the device is exposed and what uncertainty the ordinary method can support. When other components dominate, spending more on the display alone may not improve the decision.

Repeatability describes consistency under stated conditions

Repeatability is not whatever variation appeared while watching a screen. It refers to closeness of results obtained under appropriately comparable conditions. A small check might use the same device, method, reference, operator and configuration over a short interval. State how readings were repeated. A row obtained by pressing save several times on one buffered number differs from independently allowing a probe to settle and reading it again. A protocol designed to test the whole placement routine may deliberately remove and replace the probe between readings. That challenges a larger portion of the actual method.

A stable offset can coexist with close repeats. Both tips of a measuring tape may be placed consistently while a mark used as the zero is wrong. A thermometer can consistently report high because its reference adjustment is off, because its housing receives more heat than the surrounding air, or because it is always placed against a warmer surface. A tight series by itself cannot distinguish those scenarios. If a measurement is used to compare spaces, check whether the mechanism is shared and whether it cancels in that comparison. Do not assume an offset observed once cancels in every setup.

The opposite combination is possible as well: readings may cluster loosely around a reference on average. That can limit decisions from one observation even if the mean is close in the small test. More repeats can improve an estimate of an average under suitable assumptions, including a stable process and appropriate dependence structure. Closely spaced logger updates may not be independent replicates because the instrument response, environmental conditions and software smoothing can tie neighboring values together. A thousand values from one moving signal are not equivalent to a thousand independent checks.

Repeatability under one arrangement does not establish broader reproducibility. Different operators can push a soil probe to slightly different depths or allow different settling times. The same sensor could be remounted in a shade shield with a different orientation. A procedure might look exceptionally stable only because one experienced worker knows when to disregard an initial indication. If other people will use it, document the rule and include comparisons that capture relevant differences among operators, days, instruments or locations. The NIST/SEMATECH Measurement Process Characterization chapter provides the broader framework for examining stability, calibration and variation in measurement processes.

Accuracy is a comparison, not praise for a number

Accuracy concerns closeness to the quantity the procedure aims to measure. For a check with an appropriate reference, the observed indication difference is a practical quantity to report. A stable pattern of high readings can be evidence of bias relative to that reference in the tested conditions. A single difference can also arise from placement, chance variation or a limitation of the reference. State the evidence before expanding the explanation. “This probe was 0.9 degrees high relative to the check under a stated setup” is stronger and narrower than “the thermometer is inaccurate.”

Choose a reference suitable for the task. If a garden team uses one familiar thermometer to compare several new ones, the result is an agreement check unless the familiar instrument has an adequately documented assignment for the relevant conditions. Two devices may agree because they share similar limitations. Even a carefully calibrated reference must be interpreted with its uncertainty and method. Comparing a temperature reported by a weather station kilometers away with a potting bench reading is usually not a device check at all: the two sensors may be measuring genuinely different environments.

Check more than one point if you need to make claims about a range. Near 20 degrees Celsius two probes may differ by little, while near a colder or warmer decision point their differences may become more important. An observed deviation can depend on the temperature rather than being a constant offset. There can also be time and handling effects. A slowly responding probe compared just after a sharp change may appear biased in one direction after warming and another after cooling. Until a method distinguishes slow response from a stable offset, the result does not justify correcting every value by the same amount.

Use care with the language of error. In the example the chart explicitly treats the reference as an assigned comparison value. In a real project the exact temperature is usually not known independently with zero uncertainty, and the measurement itself may change if the setup changes. Report indication minus reference and describe how the reference was obtained. A practical check can identify a discrepancy large enough to change an operational decision without pretending to know a perfectly exact true temperature. The data become more usable when the setup and limitations survive in the record.

Uncertainty belongs to a result and a method

A display specification can be one contribution to uncertainty, not its total. Depending on the result, a budget might include repeat placement, reference assignment, thermal gradients, stabilization, readout resolution and instrument behavior over time. Each should be evaluated using relevant evidence and expressed in comparable terms before combining it under stated assumptions. The NIST technical note distinguishes components evaluated statistically from those evaluated using other information and explains how to report combined or expanded uncertainty. It does not say to identify the entire uncertainty with the last digit or the number of repeated readings.

Imagine the variation from uncertainty in sensor placement is bigger than the device's displayed increment. A better jig or consistent mounting method may improve the measurement more than buying an extra decimal. Alternatively the location may be well controlled, while reference uncertainty makes small changes difficult to evaluate. Improving the reference could then be worth more. A written budget helps choose the next investment by matching it to a limiting component. It should not make the apparent quality of a result depend on how many components someone decided to omit.

Some components can be shared among many records. A known uncertain reference assignment used for two instruments may affect both comparisons, and a common scale error may not average away when many observations are collected. Other components, such as repeat placement differences under suitable conditions, may behave differently when estimated with repeated work. If a study only needs a paired difference, some shared effects may be less important, but demonstrate which are plausibly shared and which are not. Two devices in different housings beside different plants will not necessarily share the same exposure-related effect just because both display Celsius.

Do not treat an unresolved bias as though fabricating a symmetric interval around the reported value solves it. Sometimes the right choice is to investigate a known direction and restrict use until it is understood. Where a validated correction is justified, record its range and the uncertainty associated with applying it. If the correction came from a comparison at one temperature, its use elsewhere needs support. Where no justified correction is available, show raw values as indications and attach the limitations rather than claiming adjusted temperatures of unsupported precision.

Formal uncertainty language should match what was done. An interval built from reading-to-reading scatter describes only that scatter under its assumptions. An instrument's published tolerance under a defined lab condition is not the same as the uncertainty of an installed garden measurement. A stitched range assembled by guessing plausible extremes has still another meaning. If the project uses an expanded uncertainty, give enough context to identify what components and coverage factor were used. If the experiment is exploratory and cannot support that calculation, clearly describe the observed differences and important unresolved influences instead of dressing them as formal coverage.

Define the garden temperature before checking the device

The phrase garden temperature can refer to many different quantities. Someone tracking weather might want shaded air temperature at a documented height and exposure. A seedling study might want soil temperature at sowing depth. A comparison of drought stress might be interested in leaf surface temperature, which requires a distinct method and interpretation. Using the same pair of instruments does not make these quantities exchangeable. Before diagnosing instrument bias, check that the proposed comparison actually targets the same quantity, place and time.

To measure an air quantity, surfaces and exposure can matter. A thermometer near a warm wall or leaf may not sample the same air as a probe in a more open or differently shielded space. Direct sunlight on a housing can influence what that sensor reports relative to a reference designed for a different exposure. Air movement may matter to how the instrument exchanges heat. Treat these as influences worth testing in a garden design, not as an excuse to discount every unexpected observation. Write down whether a discrepancy is measured against an appropriate reference or simply noticed between two locations.

For a soil quantity, contact and depth deserve special attention. Two probes placed at different depths can find real differences, particularly during periods of rapid warming or cooling near the surface. A pocket around the probe can change its thermal contact. If you intend buried probes to represent the same layer, fix a depth convention, mark how the sensing portion is located, and avoid assuming that equal insertion lengths always place sensors identically when their shapes differ. Preserve the observed positioning differences if a perfectly matched installation is not possible.

Averaging period can also change the quantity. A display may show an instantaneous-looking value from a filtered stream while a logger exports a ten-minute mean. A maximum recorded at some time after a sharp change does not necessarily represent the same phenomenon as a maximum of averaged readings. Compare like with like. For a daily garden summary record timestamp conventions, sampling cadence and aggregation method. A thermometer evaluated steadily against a reference can be useful without being fast enough to track a brief transient under field conditions.

A mismatch is not always a problem to erase. If the study intentionally compares near-leaf and open-air conditions, the differences matter because the placements differ. Separate the device check from the ecological comparison by using a common check condition initially and documenting the distinct placement during the trial. That gives the team evidence to distinguish a sensor difference from an environmental one without expecting every legitimate temperature reading in the garden to match.

Make a same-condition comparison worth the effort

Before collecting any values, name the decision the check will inform. A team deciding which thermometer can distinguish two propagation benches should test something close to the temperatures and exposures of those benches. A crew seeking a reliable outdoor ambient record should care about the installation and response under the intended shelter and airflow. The check should represent the working range and mode, not whichever scenario will make both devices agree immediately. A very narrow test is acceptable if the report clearly limits the conclusions to it.

If using a controlled environment, follow the applicable procedures and device limits rather than improvising an exposure that could damage a unit or change what it measures. Some sensors are designed for immersion, while others are not. Verify that the ideal of placing sensing elements at comparable points is physically achievable with the instruments at hand. An arrangement can be uniform enough for a broad screening decision and still be inadequate for a comparison at hundredths of a degree. Assess that difference before using the finest last digit as a pass threshold.

Define how and when to read each device. If one takes longer to settle, wait until the specified condition is reached before recording a stable reference check. If the goal is to test response time, that transient variation is useful and should be deliberately observed as a different test. Many contradictions disappear when steady-state comparisons and response comparisons are separated. A sensor that eventually reaches the right level can still miss the peak of a short heating event; a fast indication can be wrong in level even when its trend rapidly follows the environment.

An orientation or position swap is sometimes informative. Suppose the first device always occupies the left half of a reference setup and always reads high. If a repeat with appropriate stabilization moves that device to the right half, observe whether the discrepancy follows the device or the location. That will not isolate every possible interaction. Still, it can help distinguish an initial location difference from a repeatable device difference under the new test. If moving the probes changes the surrounding conditions, document that instead of claiming the design isolated one factor perfectly.

Try a small design with stable conditions at two relevant levels, more than one repeat placement, and a plan for the order. If all warmer checks are run later in the day or after a device wakes from sleep, the temperature level and session order may be difficult to separate. Repeating a level later or reversing order in a fresh run may reveal a time-related change. The working goal is not to build an elaborate calibration laboratory in a shed. It is to make a few observations carry more information than a collection of unlabeled side-by-side photographs.

Record actual time, method version, device and reference IDs, unit, assigned comparison value, placement and indication. Keep an exception field for condensate, a moved support or an incomplete warm-up. If temperature varies and readings could not be made together, preserve their time difference and the limits this introduces. A complete report should distinguish raw indications from computed paired differences. Do not overwrite a number because you later decide it belongs to a settling period. Mark its status and keep the main summary restricted to the intended test conditions.

Choose the reporting increment intentionally

A project may store raw indicated precision yet publish rounded results. Those practices are compatible. The raw reading allows checks of the conversion and processing. The reported rounding tells the reader what comparisons the project can support without misleading precision. If readers will use the numbers to compare subtle differences, it is often useful to include an uncertainty statement, quality note or explicit limit so the displayed digits are not mistaken for analytical confidence.

Consider reporting the invented mean indication of 20.855 degrees Celsius. That is the exact arithmetic result for the shown fine-display inputs. A second decimal was already the display increment; a third emerged from taking a mean of two displayed readings. The calculation does not turn the third decimal into independent sensor detail. If the appropriate report is a rough comparison, a phrase such as “about 0.9 degrees high in two comparisons” may communicate the evidence more faithfully than presenting 0.855 as a transferable device specification.

When a team does retain more decimals in calculations, avoid rounding at multiple intermediate steps unless the method requires it. Rounding each recorded difference then averaging may not produce the same number as averaging first and rounding for display. Use the same policy across devices so a coarser representation does not create an artificial impression of clustering. In the accessible chart table retain the values as authored for the example, including the visibly different increments. In narrative, distinguish computed examples from physical findings and explain why additional digits may appear in arithmetic.

Treat conversions and plots with the same discipline. Equivalent temperatures expressed in a different unit can acquire long decimal tails. A charting library can put six digits in hover text despite short labels on the main axes. If the displayed precision implies a level of confidence you did not evaluate, adjust it or clarify the distinction right there. Conversely, do not discard useful distinctions by aggressively rounding the records and then trying to infer a trend smaller than the lost increment. Keep the archival and reader-oriented views deliberately connected.

Plan for decisions close to a threshold

For a typical garden trial it may be sufficient to describe a broad pattern: one bed warmed several degrees over the morning while another changed little. A threshold question is less forgiving. If a team plans to move seedlings when a temperature reaches a specified point, what happens when the thermometer hovers very close to it? A finer screen might make the crosses look definite while the actual method leaves appreciable uncertainty. The project can set a review interval, take additional suitable readings or use an independent check when decisions are close. Do not invent a universal temperature threshold for all plants or installations.

Think through the consequences of being wrong on each side before choosing a tolerance. An early action and a delayed action may not cost the same. A trend warning can sometimes tolerate more error because it does not directly trigger work. A controlled comparison that will inform a later public claim may need a stronger method even if field operations can proceed under a simpler rule. Document which records are suitable for which decision rather than treating a screen's precision as permission to use it everywhere.

Near a boundary, repeat readings help only to the extent they address the relevant uncertainty. Waiting for an unstable display to show a convenient number defeats the purpose. Follow a stated rule for when readings are averaged, repeated or discarded. If the main concern is a location-dependent bias, repeating the same procedure at the same potentially unsuitable location may not resolve anything. An independent method, a more representative mounting arrangement or an appropriately validated reference may be a better use of time.

You can rehearse the policy with a fictional pair of indications just below and above a decision line. Ask whether both would lead to the same action under a review band, what evidence would settle a close case, and how an inability to obtain better evidence would be recorded. A clear answer helps the group resist after-the-fact changes. Keep the hypothetical numbers clearly marked as such. The exercise tests the policy, not how much confidence the team can squeeze from a convenient example.

Avoid misleading comparisons when replacing a device

A new logger often arrives halfway through a season. If its output starts higher or lower than the old pattern, the first question is whether the environment changed, the measurement changed, or both. A period of overlap under comparable placement gives more evidence than one old-afternoon reading followed by one new-morning reading. Retain the identities of both devices and changes to exposure, data interval and rounding. If there was no overlap, a visible break in the record is often more honest than an uninterrupted line suggesting nothing changed.

Do not blend two devices by adding a convenience correction derived from one moment. A correction that aligns both readings near 20 degrees Celsius may not align their colder or warmer behavior. If one device is slower, a comparison of daily extremes may reveal a difference absent at a steady reference. Make the comparison appropriate to the quantity you plan to publish. If the only supported match is a broad midrange pattern, restrict the continuity claim to that. The new device may be perfectly useful for future measurements while the old and new records remain difficult to combine.

A useful handoff includes a table listing the dates of parallel operation, the range actually observed, the check arrangement and the summaries that can be compared. Record known artifacts instead of trying to make them invisible. For a long project, the eventual question may no longer be the one the device was purchased for. Someone might investigate a cold morning weeks later or compare gradients between locations. Knowing which device produced which values and which comparisons were established will matter more than whether the plotted colors were consistent.

The reverse mistake is discarding every old record simply because the new screen has more digits. A historically coarse series may still represent broad trends well within its documented method. Compare by the level of inference it supports. An equipment replacement can improve some forms of observation and do little for others. Separate an editorial decision to show a smooth comparison from a measurement decision about what is genuinely comparable; those decisions should not automatically produce the same display.

Make the purchase question come last

When reviewing candidate instruments, ask for the operating range, allowable error or performance description under specified conditions, response behavior, display and stored resolution, calibration options and uncertainty information appropriate to the use. Read the conditions and exclusions attached to specifications. A number reported for the sensing element alone is not necessarily the performance of an installed device with all influences included. A specification evaluated in one stable environment may not address sun exposure, weatherproof installations or thermal contact in the ground. Separate documented manufacturer information from what your own comparison establishes.

Also consider whether the device makes a sound method easier. A clear export with reliable time information can preserve more investigative value than a higher-resolution display with no traceable records. A consistent mounting point can reduce placement error. A replaceable cable can simplify service if replacement preserves the relevant calibration, or require a fresh check if it changes the measurement path. A purchase made for field work should be judged partly on whether different operators can reproduce the setup and whether a fault leaves enough evidence to be noticed.

You do not have to prove that one thermometer is universally best. Choose the instrument and method that answer the intended question with an appropriately documented result. A brief note can state the quantity, setup, reference, observed deviations, repeatability scope and unresolved limits. If two tools are equally adequate for a broad purpose, ease of use may reasonably decide. If neither has demonstrated what the decision requires, do not buy confidence by selecting the one that produces more digits in a sample file.

The next time a shifting decimal catches your attention, ask whether the measured quantity is changing, the indication is stepping, the environment around the sensor is changing, or some combination is at work. A same-condition comparison can make that question concrete. The four bars make one small point visible: fine display resolution and closeness to an assigned reference need not travel together. The richer lesson is procedural. Reliability comes from defining the measurand, testing it under appropriately comparable conditions and reporting the limits of what those tests actually show.

A practical example: two benches and one reference

Imagine a volunteer team comparing overnight air temperatures at two propagation benches. One is beside the greenhouse door and the other near the back wall. In a preliminary setup each bench gets a different logger. The first preliminary report ranks the benches by tenths of a degree, but there is no record of the loggers side by side or of how each was mounted. The displays both use the same unit, yet the trial cannot distinguish a location effect from a measuring-system difference. A more useful next step is a shared reference comparison, not buying a device with an even longer display.

Document the range actually needed. A comparison made in a warm room may be useful for checking the most routine condition but irrelevant to the coldest part of the night. Begin with a controlled and suitable comparison around the temperatures where decisions are most sensitive. Put both instruments in comparable conditions for a field overlap to see whether timing and response introduce further differences. Record stable and rapidly changing periods separately. If the pattern suggests that one logger responds later after the door opens, compare both the settled indications and time behavior. Do not use a lag associated with one period as proof of constant lag at other rates of change.

The design could then exchange instruments between the benches over separate observations, preserving position and exposure records. If the apparent warmer bench follows the instrument, that is evidence worth revisiting. If the warmer ranking follows the location after the exchange, the location hypothesis gains support, although day-to-day weather and airflow still need consideration. Replication across relevant conditions is more useful than a ranking from two favorable nights. For a quick operational estimate the evidence may suffice; for a general conclusion about the structure it may only establish an initial range of conditions.

End with a statement restricted to what was observed. A report might say that at the checked range the loggers had small, documented differences relative to the comparison, and that on comparable nights the back bench stayed warmer after accounting for placement and instrument changes. If the level of agreement is too weak to support that statement, say so instead. Keep unprocessed indications and paired comparisons so another reader can evaluate a larger effect without treating the last digits as a guarantee. This is a success even if the finer logger turns out to be the better choice: you know why, and the extra digits are no longer the explanation.

Sources

Frequently asked
What is More Digits Do Not Make a Better Garden Thermometer about?
!Two differently shaped unbranded thermometers in a shaded field calibration rack, display faces turned away. (generated illustration)
What should you know about read the chart as an invented check, not a product comparison?
The four bars compare two hypothetical indications from each thermometer against an assigned reference temperature of 20 degrees Celsius. The fine display reads 20.84 and 20.87 degrees Celsius. The coarse display reads 20.2 and 20.1. These are constructed values, not results from the NIST Technical Note 1297 linked…
What should you know about resolution describes the steps, not the temperature?
Resolution is about distinguishing a small change in an indication. An obvious feature of a digital display is its displayed increment. A display advancing from 20.84 to 20.85 in the mode used by the observer shows an increment of one hundredth of a degree Celsius in that range. It might be displaying a rounded…
What should you know about repeatability describes consistency under stated conditions?
Repeatability is not whatever variation appeared while watching a screen. It refers to closeness of results obtained under appropriately comparable conditions. A small check might use the same device, method, reference, operator and configuration over a short interval. State how readings were repeated. A row obtained…
What should you know about accuracy is a comparison, not praise for a number?
Accuracy concerns closeness to the quantity the procedure aims to measure. For a check with an appropriate reference, the observed indication difference is a practical quantity to report. A stable pattern of high readings can be evidence of bias relative to that reference in the tested conditions. A single difference…